Atomic absorption rests on a single idea. Free ground-state atoms in the light path of a hollow-cathode lamp emitting the same element's characteristic line absorb a fraction of the light proportional to their number, and Beer-Lambert converts the reading into a concentration. What changes between the four modes is the way the sample is persuaded to give up its atoms.
Flame AAS uses a burner fed with fuel and oxidant. Air-acetylene at about 2300°C atomises lead at 283.3 nm, copper at 324.7 nm, zinc at 213.9 nm, cadmium at 228.8 nm. Refractory elements need nitrous oxide-acetylene at about 2800°C (aluminium, chromium, vanadium). Flame AAS sits at ppm and is the proportionate tool for acute high-dose cases.
Graphite-furnace AAS swaps the flame for a graphite tube. A microlitre aliquot is pipetted in, the tube is heated through a programmed ramp (drying 110°C, ashing 400 to 1200°C, atomisation up to 2700°C), and the atomic vapour sits in the light path long enough for the reading. Detection drops to ppb, the regime needed for As, Pb, Cd and Hg in routine blood and urine.
Cold-vapour AAS is the mercury-specific mode. Mercury is the only metal with a measurable vapour pressure at room temperature. Stannous chloride reduces mercuric ion to elemental mercury, the vapour is swept into a quartz cell, and absorbance is read at 253.7 nm.
Hydride-generation AAS exploits the hydride chemistry of As, Sb, Se, Bi and Sn. Sodium borohydride in acid reduces the metal ion to its volatile hydride (arsine AsH3, stibine SbH3), swept by argon into a heated quartz cell at about 900°C where it dissociates to free atoms.
|
| Flame AAS (air-acetylene) | Flame at about 2300°C | Pb, Cu, Zn, Cd, Ca, Mg, Fe, Mn | µg/mL (ppm) |
| Flame AAS (N2O-acetylene) | Flame at about 2800°C | Al, Cr, V, Mo, Si, Ti, Ba |