Mass Spectrometry: Principles, Ionisation and Mass Analysers

How mass spectrometers actually work: the five blocks of any MS, ionisation sources from EI to MALDI, mass analysers from quadrupole to Orbitrap, vacuum systems, detectors, and what an Indian forensic bench at CFSL, NDTL or NIPER actually runs.

Last updated:

Share

Mass spectrometry (MS) is an analytical technique that converts neutral molecules into gas-phase ions, separates them by their mass-to-charge ratio (m/z), and counts each value to produce a stick-plot mass spectrum. The spectrum yields the molecular weight from the molecular ion peak, structural information from fragment ions, and, on high-resolution instruments, a unique molecular formula from accurate mass measurement. No other bench technique combines identification specificity, sensitivity, and throughput in the same instrument family, which is why MS sits at the top of every forensic analytical hierarchy and underpins casework from drug identification to anti-doping to microbial typing.

A mass spectrometer is, at its core, a very precise scale that weighs molecules one at a time. The sample is converted into ions, those ions are sorted by their mass-to-charge ratio (m/z), and each m/z value is counted. The output is a mass spectrum: a stick plot with m/z on the x-axis and relative abundance on the y-axis, with a tall base peak, a smaller molecular ion peak, and a forest of fragment ions in between. Read that spectrum well and you can name the compound, work out its molecular formula and even reconstruct the bonds that broke when it fragmented.

Key takeaways

  • A mass spectrometer converts neutral molecules into gas-phase ions, sorts them by mass-to-charge ratio, and counts each value to produce a stick-plot mass spectrum with a molecular ion peak and fragment ions.
  • The instrument is best understood as five interchangeable blocks: the same platform can run ESI-Orbitrap for unknown psychoactives or APCI for steroid screening depending on configuration.
  • A bench Orbitrap reads accurate mass to within 5 ppm and resolves ions less than 0.001 Da apart, while a triple quadrupole running MRM detects synthetic cathinones in urine at picogramme levels.
  • MALDI-TOF identifies a bacterial isolate in about 90 seconds, illustrating the throughput advantage MS holds over other identification techniques in a forensic workflow.
  • Indian CFSL and SFSL benches have built their case-defensible chemistry on this family of instruments since the early 2000s, with the technique sitting at the top of every forensic instrument hierarchy.

The reason MS sits at the top of every forensic instrument hierarchy is that no other technique gives the same combination of identification specificity, sensitivity and throughput. A modern bench Orbitrap reads accurate mass to within 5 ppm and resolves ions less than 0.001 Da apart. A triple quadrupole running multiple-reaction-monitoring (MRM) detects a synthetic cathinone in urine at picogramme levels. A MALDI-TOF identifies a bacterial isolate in 90 seconds. The CFSL, SFSL and anti-doping benches in India have built their case-defensible chemistry on this single family of instruments since the early 2000s.

The mass spectrometer is not one instrument; it is a kit of five interchangeable blocks. The same Thermo Q-Exactive that runs ESI-Orbitrap for unknown psychoactives can be reconfigured with APCI for steroid screening or with a different inlet for direct-infusion work. Understanding MS means understanding the five blocks, because every real instrument is an opinionated combination of them.

By the end of this topic you will be able to:

  • Describe the function of each of the five instrument blocks (inlet, ion source, mass analyser, detector, vacuum system) and explain why swapping any block changes the instrument's analytical capability.
  • Interpret a 70 eV EI mass spectrum by identifying the molecular ion, base peak, and key fragment ions, and apply common neutral-loss values to infer structural features.
  • Compare hard and soft ionisation sources (EI, CI, ESI, APCI, MALDI) and select the appropriate source for a given analyte class and matrix.
  • Distinguish the six main mass analyser families by resolution, mass range, and forensic application, and explain why triple-quadrupole MRM and Orbitrap HRAM address different casework needs.
  • Explain how accurate mass measurement at 5 ppm constrains a molecular formula and why this capability is essential for novel psychoactive substance (NPS) identification in the absence of a library reference.
Key terms
Mass-to-charge ratio (m/z)
The fundamental MS measurement: the mass of an ion divided by its charge number. A singly charged ion of mass 250 Da appears at m/z 250; a doubly charged ion of the same mass appears at m/z 125. The mass analyser sorts ions by this ratio, not by mass alone.
Molecular ion (M+•)
The intact ionised molecule with no fragmentation, formed by removing a single electron from the parent. Its m/z gives the molecular weight of the analyte directly. Hard ionisation often destroys it; soft ionisation preserves it.
Base peak
The most intense peak in a mass spectrum, normalised to 100 percent relative abundance. All other peaks in the spectrum are reported as a percentage of the base peak. It is rarely the molecular ion under EI conditions.
Resolution
The ability of a mass analyser to distinguish two ions of nearly equal m/z, defined as m/Δm at a given peak. A unit-resolution quadrupole separates m/z 500 from 501; an Orbitrap at resolution 100,000 separates 500.000 from 500.005.
Accurate mass
A mass measured to four or more decimal places, sufficient to compute a unique molecular formula. High-resolution accurate-mass (HRAM) instruments (Q-TOF, Orbitrap, FT-ICR) deliver mass accuracy below 5 ppm, which is the working threshold for formula assignment.
Tandem MS (MS/MS)
Two stages of mass selection separated by a fragmentation step. The first analyser picks the precursor ion, a collision cell breaks it apart, and the second analyser scans the products. Triple quadrupole, Q-TOF and ion-trap instruments all do this in different ways.

The MS principle and the five blocks of any instrument

Every mass spectrometer ever built does the same three things in sequence. It takes neutral molecules out of the sample and turns them into gas-phase ions. It separates those ions by m/z in the gas phase. It detects each ion at the analyser's exit and counts how many of each m/z hit the detector. The output is the mass spectrum.

The catch is that the gas-phase ion is fragile. It cannot survive a collision with an air molecule, it cannot drift through a tube of finite length without scattering, and it cannot be created in the first place if the analyser is sitting at atmospheric pressure. So every mass spectrometer is built around five engineering blocks that solve those problems together.

BlockFunctionTypical componentsForensic example
Sample inletGet the analyte into the ion source as a clean, gas-phase or droplet streamGC capillary interface, LC-ESI probe, MALDI plate, direct probeGC inlet from a fire-debris extract; LC-ESI from a urine drug screen
Ion sourceConvert neutral analyte into gas-phase ions of known chargeEI filament at 70 eV, ESI needle at 3 to 5 kV, MALDI N2 laser at 337 nmEI for a cocaine GC-MS run; ESI for an LC-MS/MS opioid panel
Mass analyserSeparate ions in space or time by their m/zQuadrupole rods, ion trap, TOF flight tube, Orbitrap central electrodeTriple quadrupole MRM; Q-TOF accurate mass for an unknown NPS
DetectorCount each ion at the analyser exit and convert to a measurable currentElectron multiplier, microchannel plate, image current pickupEM on a quadrupole bench; MCP on a TOF flight tube
Vacuum systemKeep the analyser at low enough pressure that ions reach the detector without scatteringRotary roughing pump plus turbomolecular pump, sometimes ion getter10^-5 torr in a quadrupole; 10^-9 torr in an Orbitrap
Left-to-right block diagram of a bench GC-MS instrument. Sample enters via a capillary column, passes through the heated EI i
Left-to-right block diagram of a bench GC-MS instrument. Sample enters via a capillary column, passes through the heated EI ion source, is sorted by the quadrupole mass analyser, and is counted at the electron multiplier detector. The turbomolecular pump maintains vacuum throughout. All connections show the direction of ion travel.

A sixth block, the data system, sits outside the vacuum and runs the instrument: it sets scan parameters, collects the digitised detector current, deconvolutes the spectrum and runs the library search. Modern instruments fold the data system into a workstation PC running vendor software (Thermo Xcalibur, Agilent MassHunter, Waters MassLynx, Bruker Compass), but conceptually it is the seventh leg of the stool.

The reason the kit-of-blocks framing matters is that swapping any one block changes the instrument's personality. A quadrupole with EI is a GC-MS for volatile drugs. The same quadrupole with ESI is an LC-MS for biological-matrix work. A TOF analyser with MALDI is a proteomics tool; the same TOF coupled to a quadrupole and ESI is a bench HRAM screen for unknowns. Real-world instrument acquisition decisions are essentially decisions about which blocks to combine.

Reading a mass spectrum

A mass spectrum looks deceptively simple. The x-axis is m/z, usually running from low to high left to right, with units in unified atomic mass units (u or Da). The y-axis is relative abundance, normalised so the tallest peak in the spectrum is set to 100 percent and everything else is reported as a percentage of that peak.

Three peak types do most of the work. The molecular ion (M+•) is the intact ionised parent and gives the molecular weight directly. The base peak is the most intense ion in the spectrum, set by definition to 100 percent. The fragment ions are the smaller pieces produced when the molecular ion breaks apart, and the gaps between them encode which bonds broke.

Toluene (C7H8, MW 92) under 70 eV EI gives a molecular ion at m/z 92 (modest intensity), a base peak at m/z 91 from loss of one hydrogen to form the very stable tropylium cation, and smaller fragments at m/z 65 (loss of acetylene from tropylium) and m/z 39. The pattern is so characteristic that a NIST library hit returns toluene at the top of the list with a match score above 950 out of 1000.

The gaps between adjacent peaks are the second piece of structural data. A loss of 15 Da from the molecular ion almost always indicates a methyl radical leaving (CH3, mass 15). A loss of 17 indicates hydroxyl (OH). A loss of 18 indicates water. A loss of 28 is one of three things (CO, N2, or C2H4) and the chemistry of the parent decides which. A loss of 29 is most often CHO from an aldehyde or ester. A loss of 43 is acetyl or propyl. Recognising common neutral losses makes spectral interpretation systematic rather than speculative.

Loss (Da)Likely neutralStructural inference
15CH3Methyl group on the parent (almost any methylated organic)
17OHCarboxylic acid or alcohol (less commonly NH3 in amides)
18H2OAlcohol, carboxylic acid, sugar, hydrate
28CO, N2 or C2H4Aldehyde-ketone (CO), N-N bond (N2), ethyl rearrangement (C2H4)
29CHO or C2H5Aldehyde alpha cleavage; ethyl loss from an ethyl ester
31OCH3Methyl ester or methoxy substituent
43CH3CO or C3H7Acetyl from a methyl ketone; propyl loss from an alkyl chain
45COOH or OC2H5Carboxylic acid; ethyl ester
Illustrative 70 eV EI mass spectrum of methamphetamine (MW 149). The base peak at m/z 58 (α-cleavage, CH₂=NHCH₃⁺) and the mol
Illustrative 70 eV EI mass spectrum of methamphetamine (MW 149). The base peak at m/z 58 (α-cleavage, CH₂=NHCH₃⁺) and the molecular ion at m/z 149 are labelled, with key fragment ions at m/z 91 (tropylium) and m/z 119. Peak heights are representative of published reference spectra for this compound class; this figure is illustrative only.

The molecular ion is not always present. Some compounds, especially long-chain alcohols, branched alkanes and certain alkaloids, fragment so extensively at 70 eV EI that no M+• survives. The chemist then runs a softer ionisation (CI with methane reagent gas) to coax the M+1 peak out. The combined hard-soft approach gives both the molecular weight from CI and the fragment pattern from EI.

Ionisation sources from EI to MALDI

The ion source is where the sample stops being a neutral molecule and starts being a measurable ion. There is no single best ionisation technique because the right one depends on what the analyte is. A volatile drug standard wants electron impact; a peptide wants electrospray; a bacterial protein wants MALDI. Pick wrong and either no ion is formed or the wrong ion dominates.

Two classifications cut across every ionisation source. Hard ionisation deposits enough excess energy that the molecular ion fragments extensively, giving a rich spectrum that can be searched against a library. Soft ionisation deposits very little excess energy, so the molecular ion (or a protonated or sodiated adduct) survives intact and dominates the spectrum, but no fragment information is produced unless an extra MS/MS step is added.

SourceHard or softSuitable analyteTypical instrument coupling
EI (Electron Impact, 70 eV)HardVolatile, thermally stable, non-polar to mid-polarGC-MS (the standard); library searchable against NIST
CI (Chemical Ionisation, methane reagent gas)SoftSame as EI, when M+• is suppressedGC-CI-MS for molecular-weight confirmation
ESI (Electrospray Ionisation)SoftPolar, ionisable, thermally labile (drugs, peptides, proteins)LC-ESI-MS/MS (the standard for biological-matrix drug work)
APCI (Atmospheric Pressure Chemical Ionisation)SoftSemi-volatile, less polar than ESI sweet spotLC-APCI-MS for steroids, lipids, pesticides
APPI (Atmospheric Pressure Photo Ionisation)SoftNon-polar that resist ESI and APCI (PAHs, sterols)LC-APPI-MS for environmental and petroleum work
MALDI (Matrix-Assisted Laser Desorption Ionisation)SoftProteins, peptides, polymers, intact bacteriaMALDI-TOF for proteomics and microbial identification
DART (Direct Analysis in Real Time)SoftSurface residues, no preparationAmbient open-air MS for rapid screening
SIMS (Secondary Ion MS)HardSurface composition, semiconductorsSurface-imaging MS in materials labs

EI is the workhorse and worth understanding in detail. The sample is vaporised into a 70 eV electron beam from a heated rhenium or tungsten filament. The collision strips one electron from the analyte to form a radical cation (M+•) carrying an excess of about 60 to 65 eV of internal energy, which is more than enough to break almost any single bond. The result is a reproducible fragmentation pattern that depends only on the analyte's structure, not on instrument vendor or operating year. This reproducibility is what makes the NIST 2023 EI library, with its 394,000 plus reference spectra, the most powerful identification tool in forensic chemistry. Drop a clean EI spectrum into a NIST search and a match score above 800 out of 1000 is, for routine drug work, identification.

ESI is the workhorse on the LC side. A solution of the analyte in a polar solvent is sprayed through a fine capillary held at 3 to 5 kV against a counter-electrode. The high field strips charge from the spray, which evaporates progressively in a heated curtain gas, and ions are emitted into the vacuum through a heated capillary. The result is a soft, near-gentle protonation that gives [M+H]+ in positive mode (most drugs) and [M-H]- in negative mode (carboxylic acids, sulphates, phosphates). Sodium adducts ([M+Na]+) and ammonium adducts ([M+NH4]+) are common and need to be recognised on the spectrum, not mistaken for unrelated ions.

MALDI is mechanistically distinct from the atmospheric-pressure sources. The analyte is co-crystallised with a UV-absorbing matrix (sinapinic acid, alpha-cyano-4-hydroxycinnamic acid, dihydroxybenzoic acid) on a stainless steel plate. A pulsed N2 laser at 337 nm fires at the spot, the matrix absorbs the laser energy and ablates a plume that carries soft-ionised analyte. Almost exclusively coupled to TOF analysers (because the laser is pulsed and TOF is naturally a pulsed-ion technique), MALDI-TOF is the standard for protein, peptide and intact-microbe identification, and it has replaced biochemical assays for bacterial typing in clinical microbiology and food forensics.

Mass analysers from quadrupole to Orbitrap

The mass analyser is where the actual sorting happens. There are six families in routine use and the choice between them shapes the entire instrument's sensitivity, resolution, mass range, scan speed and price. A bench drug-screen quadrupole costs about INR 40 lakh; an Orbitrap with the same throughput but 100 times the resolution costs INR 3 to 5 crore. The price is not arbitrary; it reflects the analyser physics.

Cross-section through a quadrupole mass analyser looking along the ion beam axis. Four rods (two pairs, shown in blue and gre
Cross-section through a quadrupole mass analyser looking along the ion beam axis. Four rods (two pairs, shown in blue and grey) carry opposing RF and DC voltages. Only an ion with the correct m/z follows a stable helical trajectory through the centre and exits to the detector; all others spiral into the rods and are lost.

The quadrupole is the most common analyser in forensic chemistry. Four parallel cylindrical rods are arranged with opposite pairs electrically connected. A combination of a fixed DC voltage and a radio-frequency AC voltage is applied to the rod pairs, which creates a quadrupolar field along the central axis. Only ions with a specific m/z (set by the RF/DC ratio) follow stable trajectories and reach the detector at the far end; all other ions hit the rods and are lost. Scanning the RF and DC voltages together sweeps which m/z passes through, building the spectrum. Quadrupoles handle m/z 1 to about 3000, run at unit resolution, are cheap, robust and fast, and are the default for GC-MS and routine LC-MS. A triple quadrupole (QqQ) puts a collision cell between two analytical quadrupoles and runs MRM transitions for the picogramme-sensitivity work that anti-doping and toxicology depend on.

The ion trap holds ions inside a 3D electric field rather than letting them transit. A 3D Paul trap (LCQ on a Thermo) uses a ring electrode and two end-cap electrodes to confine ions in a closed orbit; a linear ion trap (LIT, the Thermo LTQ) uses an extended quadrupole with end electrodes. Mass-selective ejection scans the trap by ramping the RF amplitude until ions of successively higher m/z become unstable and fall out of the trap into the detector. The ion trap's signature trick is MSn: trap a precursor, fragment it in-trap, trap a chosen fragment, fragment it again, and so on for several cycles. This sequential fragmentation is unmatched for structural elucidation of unknowns.

The time-of-flight (TOF) analyser operates on a straightforward principle. A pulse of ions is accelerated through a fixed potential difference (usually 20 to 30 kV), giving every ion the same kinetic energy. The ions then drift down a field-free tube; lighter ions arrive first and heavier ions later, and the arrival time encodes m/z. A reflectron at the end of the tube reverses the ions back along a slightly longer path, correcting for kinetic-energy spread and pushing resolution from a few thousand to about 50,000. TOFs handle very large masses (well over 100,000 Da, with no upper limit in principle) and are the natural partner for MALDI's pulsed-laser ionisation.

The quadrupole-TOF (Q-TOF) marries a quadrupole front-end with a TOF analyser. The quadrupole acts either as a wide-bandpass ion guide (TOF gets every ion) or as a precursor selector for tandem MS (TOF measures only fragments of the chosen precursor). The result is a high-resolution accurate-mass (HRAM) instrument that gives both qualitative breadth and structural depth, which is why the Bruker Compact, Waters Synapt and Sciex TripleTOF are the workhorses of unknown-NPS screening.

The Orbitrap is the newer high-end design. Ions are injected tangentially into a barrel-shaped outer electrode wrapped around a central spindle electrode, and they orbit the spindle while oscillating along its axis. The axial oscillation frequency depends only on m/z, and an image-current pickup at the outer electrode senses every oscillating ion simultaneously. Fourier-transforming the resulting time-domain trace gives a frequency spectrum, which converts to an m/z spectrum at resolution above 100,000 and mass accuracy below 5 ppm. The Thermo Q-Exactive, Exploris and Orbitrap Astral series are the dominant HRAM instruments for forensic toxicology and metabolomics.

The magnetic sector is the classical analyser, where a magnetic field bends the ion path with a radius proportional to the square root of m/z. Combined with an electric sector for energy filtering (a double-focusing instrument), it delivers high resolution and very accurate mass, but the instrument is expensive, heavy and slow to scan. Modern use is largely confined to isotope-ratio MS (NDTL Delhi runs GC-C-IRMS for steroid origin analysis) and a few specialised laboratories.

The FT-ICR (Fourier-Transform Ion Cyclotron Resonance) is the highest-resolution instrument ever built. Ions trapped in a strong magnetic field (typically 7 to 21 tesla) cyclotron at a frequency inversely proportional to m/z. Image-current detection followed by Fourier transform gives resolution above one million and mass accuracy at the sub-ppm level, but the instrument costs upwards of INR 10 crore and is restricted to petroleomics and intact-protein characterisation in academic centres.

AnalyserResolution (m/Δm)Mass rangeTypical role and Indian deployment
Quadrupole (single)About 1000 (unit)1 to 3000 DaGC-MS and LC-MS routine; Agilent 5977 at FSL Madhuban
Triple quadrupole (QqQ)About 1000 (unit)1 to 3000 DaMRM picogramme quantitation; Waters Xevo TQ-S at NDTL Delhi
Ion trap (3D Paul or LIT)About 1000 to 400050 to 4000 DaMSn structural elucidation; LCQ and LTQ in academic labs
TOF (with reflectron)About 20,000 to 50,000Up to 500,000 DaMALDI-TOF for proteins, polymers; clinical microbiology benches
Q-TOFAbout 30,000 to 50,0001 to 40,000 DaUnknown NPS screening; Bruker Compact at NIPER Mohali
OrbitrapAbove 100,00050 to 6000 DaHRAM forensic toxicology; Thermo Q-Exactive at CFSL Hyderabad
Magnetic sector (double-focusing)About 60,0001 to 10,000 DaIsotope-ratio MS; GC-C-IRMS at NDTL Delhi
FT-ICRAbove 1,000,00050 to 10,000 DaPetroleomics and academic intact-protein; restricted to NCBS class facilities

Detectors and the vacuum system

The detector at the analyser exit converts each arriving ion into a measurable electrical pulse. The dominant choice for quadrupole and ion-trap instruments is the electron multiplier (EM), a discrete-dynode or continuous-channel device in which an arriving ion strikes the first dynode and ejects 2 to 4 secondary electrons; each of those strikes the next dynode and ejects more, and the cascade through 12 to 20 stages amplifies a single ion arrival into a measurable pulse of about a million electrons. The gain is roughly 10^6 and the response is fast enough for unit-mass scans at thousands of m/z per second.

The microchannel plate (MCP) is a plate version of the EM, made of millions of parallel glass capillaries, each lined with a secondary-emitting coating. Ions striking any one capillary trigger a local cascade. MCPs preserve spatial information across the plate, which is what TOF analysers need (every m/z arrives at a different time at the same plate face) and what imaging-MS instruments depend on.

The Faraday cup is the simplest detector: a metal cup catches the ion and the resulting current is measured directly. There is no gain, so sensitivity is poor, but the response is exactly linear with ion abundance and is therefore used for very high-abundance ions (the molecular-leak channel of a sector instrument, or as a reference detector in isotope-ratio work).

The Orbitrap is unusual in not having an ion-collecting detector at all. Image-current detection at the outer electrode senses the moving ions inductively without absorbing them, which is also what FT-ICR does. The signal is a time-domain trace that is Fourier-transformed to give the spectrum. A photomultiplier tube paired with a scintillator is sometimes used at the orbital-trap exit for daughter-ion analysis.

Vacuum is the second non-negotiable. Ions in the gas phase have a mean free path between collisions that is inversely proportional to pressure. At atmospheric pressure (760 torr) the mean free path is about 70 nm, far shorter than any analyser. At 10^-5 torr the mean free path is roughly 5 metres, comfortably longer than a quadrupole or ion trap. At 10^-7 torr it is hundreds of metres, enough for a TOF flight tube. Below this the ions reach the detector unscattered and the m/z measurement is reproducible.

Three pump types build the vacuum together. A rotary roughing pump pulls the analyser from atmospheric pressure down to about 10^-2 torr. A turbomolecular pump takes over from there, spinning a stack of bladed rotors at 60,000 to 90,000 rpm to drag residual gas molecules out, reaching 10^-5 to 10^-9 torr. An ion getter pump or a cryopump is added on the highest-vacuum analysers (Orbitrap, FT-ICR) to push pressure down to 10^-10 torr or lower, holding it there with no moving parts.

Forensic applications and the Indian instrument map

The Indian forensic-MS landscape settles around four vendors and a handful of instrument families. The Agilent 7890 GC paired with the 5977 single-quadrupole MSD is the workhorse GC-MS for drugs, fire-debris analysis and pesticide residues at most state SFSLs. The Agilent 6470 triple quadrupole LC-MS/MS handles biological-matrix toxicology and food-residue work. CFSL Chandigarh runs both for the chemistry caseload of the northern region.

The Thermo Q-Exactive Orbitrap is the high-resolution flagship at CFSL Hyderabad, where it serves both forensic chemistry (NPS screening, unknown plant alkaloids) and forensic biology (proteomic identification of body fluids). The Q-Exactive's combination of an ESI-quadrupole front-end with the Orbitrap mass analyser delivers the 5 ppm accurate mass that converts an unknown spectrum into a unique molecular formula in seconds, which is precisely what the unknown-NPS workload demands.

The Waters Xevo TQ-S triple quadrupole is the anti-doping bench's tool of choice. NDTL Delhi runs an Xevo TQ-S MRM panel covering the WADA prohibited list across about 250 transitions, with picogramme-per-millilitre limits of detection. Paired with a GC-C-IRMS magnetic-sector instrument for the carbon-isotope test that distinguishes endogenous from exogenous testosterone, NDTL operates one of the most demanding MS workloads in the country.

The Bruker Compact Q-TOF at NIPER Mohali serves the academic and collaborative research workload, with applications spanning unknown bulk-drug characterisation, counterfeit pharmaceutical investigations and MS-imaging research. The Q-TOF's accurate mass and tandem MS depth make it the right second-line instrument when an in-house library does not return a hit.

The Sciex 6500+ QTRAP at CDFD Hyderabad blends triple-quadrupole MRM with linear-ion-trap functionality on the same platform, covering both quantitative drug analysis and qualitative metabolite identification in the same run. The QTRAP is the natural choice when a lab needs both MRM throughput and MSn depth without buying two separate instruments.

Indian labPrimary instrumentWorkflowCasework
CFSL ChandigarhAgilent 7890+5977 GC-MS, 6470 LC-MS/MSEI library search; ESI MRM panelDrugs, pesticides, fire debris, biological-matrix toxicology
CFSL HyderabadThermo Q-Exactive OrbitrapESI HRAMNPS screening, unknown plant toxins, proteomic body-fluid identification
FSL Sector 14 MadhubanAgilent GC-MSEI library searchDrugs and pesticide residues
NDTL DelhiWaters Xevo TQ-S MS/MS, GC-C-IRMS magnetic sectorMRM and isotope-ratioWADA anti-doping panel; endogenous vs exogenous testosterone
NIPER MohaliBruker Compact Q-TOFESI accurate mass and MS/MSCounterfeit pharmaceuticals, unknown bulk-drug characterisation
NCBS BangaloreThermo Orbitrap seriesESI HRAM and proteomicsAcademic and collaborative work, intact-protein, metabolomics
CDFD HyderabadSciex 6500+ QTRAPMRM with LIT MSnQuantitative toxicology with structural confirmation

The casework patterns sort cleanly by instrument family. GC-MS handles every volatile and thermally stable analyte: classical drugs of abuse (heroin, cocaine, methamphetamine), fire-debris work (petroleum distillates and accelerants), explosives screen (nitroaromatics), pesticide residues and environmental contaminants. LC-MS/MS handles polar, ionisable, thermally labile analytes in biological matrix: drugs in urine and blood, anti-doping screening, pesticide residues in food, mycotoxins. HRMS Q-TOF and Orbitrap handle unknowns that need accurate mass: novel psychoactive substances entering the market, unknown plant alkaloids in poisoning cases, intact protein characterisation. MALDI-TOF handles bacterial identification (food forensics, post-mortem microbiology), peptide profiling and polymer characterisation.

The analytical demands are consistent across these workflows: clean spectral interpretation (molecular ion, base peak, fragment pattern, isotope pattern), molecular formula assignment from accurate mass at 5 ppm or better, NIST library search with score-versus-overlay evaluation, and neutral-loss pattern recognition for structural inference independent of library coverage.

Practice
Question 1 of 5· 0 answered

What does the m/z value of an ion in a mass spectrum represent?

Frequently asked questions

Why is the molecular ion often missing or very small in an EI mass spectrum?
70 eV electron impact deposits about 60 to 65 eV of internal energy into the analyte, which is much more than any single bond dissociation energy (typically 3 to 5 eV). The molecular ion fragments extensively before it can reach the detector, especially for compounds with weak alpha-cleavage pathways (long-chain alcohols, branched alkanes, certain alkaloids). Switching to chemical ionisation with methane reagent gas deposits much less excess energy and gives an [M+H]+ peak that survives intact, which is the standard workaround when EI shows no molecular ion.
What is the difference between resolution and mass accuracy on a mass spectrometer?
Resolution is the analyser's ability to distinguish two ions of nearly equal m/z, expressed as m/Δm at a defined peak. A unit-resolution quadrupole separates m/z 500 from 501; an Orbitrap at resolution 100,000 separates 500.000 from 500.005. Mass accuracy is how close the measured m/z is to the true theoretical value, expressed in ppm. A high-resolution instrument can have poor mass accuracy if poorly calibrated, and a unit-resolution quadrupole can have respectable mass accuracy for nominal-mass identification. Both matter for unknown identification.
Why are MALDI and TOF almost always paired together?
MALDI is an inherently pulsed ionisation source: each laser shot generates a brief packet of ions with a defined start time. TOF needs exactly that, because the analyser's whole principle is timing the arrival of ions at the detector relative to a known start time. A continuous ionisation source (ESI, EI) would smear the start time and make TOF measurement meaningless without an additional ion-pulsing stage. The mechanical coincidence of MALDI's pulsed output and TOF's pulsed-input requirement is why MALDI-TOF is the dominant pairing for protein, peptide and intact-microbe identification.
What is multiple-reaction-monitoring (MRM) and why does the anti-doping bench rely on it?
MRM is a tandem-MS workflow on a triple quadrupole. The first quadrupole selects the precursor ion at the analyte's molecular m/z; the collision cell fragments it; the third quadrupole monitors only one or two specific product ions characteristic of that analyte. By measuring only those preset transitions and skipping the rest of the spectrum, MRM achieves picogramme-per-millilitre limits of detection on real biological matrices. NDTL Delhi runs about 250 MRM transitions covering the WADA prohibited list on every doping sample, which is the only way to screen at the required regulatory sensitivity within a 24-hour turnaround.
How is accurate mass used to determine a molecular formula from a single MS measurement?
Each combination of carbon, hydrogen, nitrogen, oxygen and other heteroatoms gives a unique exact mass that differs from any nearby combination by tens or hundreds of millidaltons. An accurate-mass measurement at 5 ppm or better is usually enough to constrain the molecular formula to a single candidate within sensible chemistry. For a small molecule of mass 200, 5 ppm is 1 mDa, easily resolving C12H12N2 (200.0950) from C11H12N2O (200.0950 with isotope correction) using accurate mass plus isotope pattern. The formula is the starting point for structural identification by MS/MS or NMR.
Why does the Indian NPS workload need an Orbitrap or Q-TOF rather than a standard GC-MS?
Novel psychoactive substances enter the market faster than NIST or the SWGDRUG monograph collection can be updated, so a library search on a quadrupole GC-MS often returns no hit or a misleading low-similarity match. An Orbitrap or Q-TOF measures accurate mass to 5 ppm and gives a unique molecular formula even with no library reference, and the MS/MS fragmentation pattern then narrows the candidate to a specific isomer. CFSL Hyderabad's Q-Exactive and NIPER Mohali's Bruker Compact handle the NPS workload precisely because they can characterise an unknown without an existing reference spectrum, which the quadrupole instruments cannot.
Is GC-MS or LC-MS/MS the right choice for drug screening in a urine sample?
LC-MS/MS is almost always the right choice for biological-matrix work. Most drugs in urine are polar metabolites (glucuronide and sulphate conjugates, hydroxylated phase I metabolites) that are non-volatile and thermally labile, ruling out direct GC-MS without derivatisation. ESI ionises them softly and the triple quadrupole MRM panel quantifies them at picogramme-per-millilitre sensitivity. GC-MS retains a role for the parent drug after enzymatic deconjugation and silyl derivatisation, but the throughput and matrix tolerance of LC-MS/MS make it the workhorse for routine urine toxicology at every modern Indian SFSL.

Test yourself on Instrumental Techniques with free, timed mocks.

Practice Instrumental Techniques questions

Found this useful? Pass it along.

Share

Spotted an error in this page? Report a correction or read our editorial standards.

Your journey to becoming a forensic professional starts here.

Practice with mock tests, learn from structured notes, and get your questions answered by a global forensic community, all in one place.