The detector at the analyser exit converts each arriving ion into a measurable electrical pulse. The dominant choice for quadrupole and ion-trap instruments is the electron multiplier (EM), a discrete-dynode or continuous-channel device in which an arriving ion strikes the first dynode and ejects 2 to 4 secondary electrons; each of those strikes the next dynode and ejects more, and the cascade through 12 to 20 stages amplifies a single ion arrival into a measurable pulse of about a million electrons. The gain is roughly 10^6 and the response is fast enough for unit-mass scans at thousands of m/z per second.
The microchannel plate (MCP) is a plate version of the EM, made of millions of parallel glass capillaries, each lined with a secondary-emitting coating. Ions striking any one capillary trigger a local cascade. MCPs preserve spatial information across the plate, which is what TOF analysers need (every m/z arrives at a different time at the same plate face) and what imaging-MS instruments depend on.
The Faraday cup is the simplest detector: a metal cup catches the ion and the resulting current is measured directly. There is no gain, so sensitivity is poor, but the response is exactly linear with ion abundance and is therefore used for very high-abundance ions (the molecular-leak channel of a sector instrument, or as a reference detector in isotope-ratio work).
The Orbitrap is unusual in not having an ion-collecting detector at all. Image-current detection at the outer electrode senses the moving ions inductively without absorbing them, which is also what FT-ICR does. The signal is a time-domain trace that is Fourier-transformed to give the spectrum. A photomultiplier tube paired with a scintillator is sometimes used at the orbital-trap exit for daughter-ion analysis.
Vacuum is the second non-negotiable. Ions in the gas phase have a mean free path between collisions that is inversely proportional to pressure. At atmospheric pressure (760 torr) the mean free path is about 70 nm, far shorter than any analyser. At 10^-5 torr the mean free path is roughly 5 metres, comfortably longer than a quadrupole or ion trap. At 10^-7 torr it is hundreds of metres, enough for a TOF flight tube. Below this the ions reach the detector unscattered and the m/z measurement is reproducible.
Three pump types build the vacuum together. A rotary roughing pump pulls the analyser from atmospheric pressure down to about 10^-2 torr. A turbomolecular pump takes over from there, spinning a stack of bladed rotors at 60,000 to 90,000 rpm to drag residual gas molecules out, reaching 10^-5 to 10^-9 torr. An ion getter pump or a cryopump is added on the highest-vacuum analysers (Orbitrap, FT-ICR) to push pressure down to 10^-10 torr or lower, holding it there with no moving parts.